找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

CX4阻挡探测校准

[复制链接]
郑振寰 发表于 2005-10-30 21:52 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

*** Submitted for Approval ***

CX4 Obstruction Detection Calibration

Last Edited 05/01/2001 08:09:31 PM

Product: Synchron CX

Chapter: CX PRO

Section: Obstruction Detection

Summary

The calibration procedure for CX4 obstruction detection and correction (ODC) is described below.

Background

The obstruction Detection and Correction (ODC) feature for CX PRO includes an automatic onboard calibration process. The calibration values are stored in nonvolatile memory and are checked against predetermined limits.

Procedure

The CX4 obstruction detection calibration procedure should be performed after installing new obstruction detection hardware. In addition, when ODC hardware is installed or new tubing is installed, a minimum of 10 internal probe rinses should be performed on the sample probe before running the calibration procedure. The primes ensure the lines are degassed sufficiently to provide a good calibration.

Access the calibration screen as follows:

From the Main Screen, select F4 Special Functions.

From the next screen, select option 6. Maintenance.

From the next screen, select option 5. CX4 Obstruction Detection Calibration.

Step

Action

The calibration procedure takes 4 measurements. Each are described below.

1

Probe measurement (no liquid) - Value 1 on the screen

This value represents the vacuum generated by the column of fluid in the sample probe. Gravity is the only force acting on the column of fluid and the gravitational pull causes a slight vacuum to be felt by the diaphragm in the obstruction assembly.

Home the CX system - Top 2 LEDs on motor driver 3 are ON.

Perform one internal probe prime

Send a command to PIC to go into calibration mode - First and third LEDs on motor driver 3 are ON.

Send a command to PIC to take the first measurement

Read PIC status byte and data byte

2

DI water measurement - Value 2 on the screen

Move the sample wheel to slot 1 cup 1 position

Aspirate an air slug

Move the sample probe and level sense DI water

Send a command to PIC to take the second measurement

Aspirate 25 uL water

Read PIC status byte and data byte

3

Sample Obstruction measurement - Value 3 on the screen

This value is calculated from the DI water measurement. It is ~15% of the DI water measurement value (value 2)

This value represents a ~0.5 uL clot.

4

Blowout measurement - Value 4 on the screen

When fluid is pushed through the probe, the diaphragm is subjected to the pressure required to push fluid out of the probe. This value will be used to determine if the probe is obstructed.

Go to wash station and inject what was aspirated in previous steps.

Aspirate DI water

Send a command to PIC to take the fourth measurement

Dispense the DI water

Read PIC status byte and data byte

5

Obstruction Detection Calibration Status Displayed on Screen

Display calibration values on the screen.

If all Pass, ask operator if want to save values (Done with Obstruction detection calibration below).

If Fail, notify operator of failure and provide option to exit calibration procedure (Abort Sequence below).

6

Done with Obstruction detection calibration

Send a command to PIC to write the calibration values to EEPROM

If PIC encounters an error in writing EEPROM, a pop-up message appears.

Read PIC status byte to confirm calibration was saved

PIC automatically performs a local software reset - Top 2 LEDs on motor driver 3 are ON.

7

Abort Sequence

If operator aborts calibration, CA Abort command is sent to PIC

PIC will perform local software reset and reload previous calibration values.

Home the CX system

Status Byte

The status byte of the PIC is read several times during the calibration. Meaning of the status byte for calibration are below, however, these values are not viewable (event with cxlogger).

04 hex = No errors, calibration data is good

14 hex = No errors, calibration data is bad

x8 hex = Error occurred, stop calibration. High nibble (x) gives specific error code.

Typical Values/Limits

The limits for each measurement are provided below. The middle value represents typical values for in-house instruments. All values are in hexadecimal (hex).

Probe Measurement (value 1): 6A<76<82 hex

DI Water Measurement (value 2): 59<65<71 hex

Sample Obstruction Measurement (value 3): 4A<56<62 hex

Blowout Measurement (value 4): 8C<98<A4 hex

Troubleshooting

The following troubleshooting tips should be useful if ODC calibration fails.

Symptom

Troubleshooting Tips

Values 1, 2 and 3 all FAIL and value 4 PASSes.

Check cable connection to motor driver 3 board.

Values 2 and 3 FAIL and values 1 and 4 PASS

Value 2 is determined while aspirating sample. If it fails, value 3 will fail. This failure mode is most likely associated with the fluid level in the cup. Check the following:

Does the cup contain fluid? Either cup was empty or insufficient fluid was in cup for procedure. Put appropriate fluid in cup and repeat procedure.

Does position 1 of the sector contain a cup? This procedure does not flag level sense errors (insufficient fluid or no fluid). Put container with appropriate fluid in position 1 of sector and repeat procedure.

Value 1 PASSes and Values 2, 3 and 4 all FAIL

Value 2 is determined while aspirating sample. If it fails, value 3 will fail. Value 4 is determined while pushing the fluid out of the probe. The limits are set with the assumption the calibration is being performed with an aqueous solution. This failure mode is most likely associated with the fluid in the cup. Check the following:

Is the fluid in the cup aqueous (DI water, saline, aqueous calibrator, etc)? If not, put appropriate fluid in cup and repeat procedure.

Vales 1, 2, 3 and 4 all FAIL

The most likely scenario for all values failing is a clogged probe or a ruptured membrane on the transducer.

The main difference between a clogged probe and a blown sensor is the magnitude of values 1and 2.

If the sensor is ruptured, all the values will be low. For one in-house system which had a ruptured CX4 ODC assembly, value 1 was 00, value 2 was 01, value 3 was F1 and value 4 was 12.

For one in-house system which had a clogged CX4 sample probe, value 1 was E0, value 2 was 07, value 3 was F8 and value 4 was 12.

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-1 06:19 , Processed in 1.311492 second(s), 32 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表