找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

BT2000生化仪通讯协议

[复制链接]
赵光 发表于 2007-12-13 19:39 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

The analyzer BT2000 PLUS allows bi-directional communication through RS 232C serial

connection with any host computer.

The particular feature of the dialog is that it is always the host computer, which initiates the

communication for either transmitting patient list or for receiving the results.

To initiate any communication the host computer will have to send to analyzer the character

STX (0x02) and expect the character ACK (0x06) as a response. At this point the host

computer will send data to the analyzer and terminate the communication by sending the

character EOT (0x04).

It is important to remember that any communication is followed by a response from the

analyzer.

It must be noted that if the parameter to be transmitted is shorter in length than the length

requirement of the communication protocol than a space must be added before or after. For

example the analysis have length 4, therefore to send a code GLI one must add a space

before or after to reach the length of 4 characters.

Patient transmission

► Start communication with sequence STX<->ACK.................

► Send patient code.................................................................. (15 characters)

► Send list type for patient insertion......................................... ("T" for Routine or "R" for STAT)

► Send type of serum............................................................... ("S" for Serum or "U" for Urine)

► Send if the patient is a clone................................................. ("Y" for Yes or "N" for No)

► Transmit position of cup........................................................ ("00" unknown)

► Send number of tests to be executed.................................... (from "01" to "99")

► Send codes of tests to be performed..................................... (4 characters)

► Send Check-Sum................................................................... (3 characters)

► Send end transmission character EOT..................................

► Wait for response from the analyzer...................................... (2 characters)

If the communication is successful then the analyzer responds with character "Y" followed

by a byte, which identifies the position where patient has been inserted. In case the

communication was unsuccessful, then the analyzer responds with "N" followed by a byte

identifying the type of error. The possible errors generated by the analyzer in response to

the invalid insertion of patient are as follows:

0x01..................... Check-Sum Error

0x02..................... Unknown Command

0x03..................... Routine/STAT field Error

0x04..................... Serum/Urine field Error

0x05..................... Clone Yes/No field Error

0x06..................... Cup position Error

0x07..................... Number of Analysis field Error

Section II Chapter 4 Serial Communication BT2000 PL Rev.0, Soft Ver. 8 Page 3 of 16

0x08..................... Wrong Number of Test

0x09..................... Position already in execution

0x0A.................... Cloning impossible

0x0B.................... Code duplicated

0x0C.................... One or more analysis not present in the analyzer

0x0D.................... One or more analysis not present in the current plate

0x0E.................... Too many analysis for the patient

For example to send a patient with code 000000000000001, serum type and with analysis

GLI, BUN and COL onto the STATS list, then one must send the following sequence of

characters (excluding initial sequence STX<->ACK):

000000000000001RSN0003GLI BUN COL 134<EOT>

Where:

000000000000001............. Patient code

R........................................ Identifies STATS list

S........................................ Identifies the type of patient (in this case: Serum)

N........................................ Identifies that the patient is not a clone

00....................................... Unknown position (the analyzer will insert the patient in a

convenient position)

03....................................... identifies the number of test to be executed.

GLI, BUN, COL................. Test codes (observe the space after each code to reach the 4

characters limit)

134..................................... Identifies the Check-Sum

<EOT>............................... This character ends communication

Results reception

There are three commands for receiving reports from the analyzer:

R.......... Reception of next available report

L.......... Reception of the last report sent (in case of reception problems)

A.......... Reception of the first available report (in case one desires to receive again all

the reports)

The commands R, L, and A require standard communication or the procedure STX<->ACK

and the character EOT to end communication.

As a response to one of these three commands the analyzer sends the requested report (if

available) or the character NAK (0x15) if there is no report to be sent. It must be borne in

mind that after a run test the reports are not immediately available for transmission as these

need validation. To do this, go to Utility menu, RS232 and enable the option "Accept result

to be sent". This operation must always be performed after each run test or groups of run

test.

There is also an additional option for performing validation operation automatically. Go to

Setup of the analyzer (Menu Utility, Setup Analyzer), go to the Serial (fourth from the left)

and enable the option "All results must be sent automatically (without validate)" at the

bottom of the page.

In case of positive response to the request for a report the analyzer transmits:

Patient code........................ 15 characters

List type............................... "T" for Routine or "R" for STATS

Sample type........................ "S" for Serum or "U" for Urine

Number of reports.............. 3 characters

For each report:

Analysis code..................... 4 characters

Result.................................. 7 characters

Check-Sum......................... 3 characters

<EOT>

The following is an example of eventual response to the data sent in "Sending a patient to

BT2000 PLUS":

000000000000001RS003GLI 000.000BUN 0010.10COL 00100.0107<EOT>

Where:

000000000000001........ Patient code

R.................................... Identifies STATS list

S.................................... Identifies the type of patient (in this case: Serum)

003................................ Identifies numbers of reports

GLI................................ First test code

000.000......................... GLI test result

BUN.............................. Second test code

0010.10......................... BUN test result

COL............................... Third test code

00100.0......................... COL test result

107................................ Identifies the Check-Sum

<EOT>.......................... This character ends communication

Calculation of Check-Sum:

This procedure calculates a Control code in accordance with the transmitted or received

data. An algebraic sum of ASCII values of all the sent characters is executed. For example

the character "A" has ASCII value 65 - 0x41.

Consequently the module 256 of the found value is executed (balance of dividing the value

by 256). This is the Check-Sum to be sent.

 


yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-24 19:58 , Processed in 2.648899 second(s), 32 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表