ATL超声诊断设备维修资料总结

[复制链接]
郑振寰 发表于 2007-1-3 13:42 | 显示全部楼层 |阅读模式

学习型号:超九、超九HDI、超四

参考资料:超九英文维修手册,超九HDI英文维修手册

型号简称:超九(简称UM9)、超九HDI(简称HDI)、超四(简称UM4)

具体型号:超九有三种: phase1 , phase2 , digital plus(DP型)

超九HDI有: UM9 HDI、 HDI.w.ESP

●UM 9


Control panel interface PCB

Control panel functions, keyboard function ,trackball function ,

footswitch function ,led function

Image memory PCB A1 Main system memory

Control panel interface PCB在触摸屏下部,键盘及触摸屏输入都连接到这块线路板上。

例如:键盘及控制面板无法输入,是这块电路板出现故障。

设备起动时,前面驱动器先工作读一号盘1秒~21秒,指示灯亮。后驱动器读2号盘

22秒~56秒,指示灯亮。

驱动器位置有两种位置分别为:在设备后盖部位或在设备整体推出后右侧位。

更换电路板注意事项:1.对照老板子芯片与新电路板芯片是否缺少。

2.注意跳基线是否相同,不同则以老电路板为准。

ACQUISITION GROUP 接收信号部

DIGITAL BEAMFORMER

探头接口板 Interconnect Multiplex PCB

n multiplexer control

n relays & 8-channel high voltage switches

n personality data interface

通道板 CHANNEL PCBS A25,A26,A27,A28,A29,A30 共六块.

图象出现干扰有可能通道板坏

n 6PCBs with 8 channels each

n variable gain receiver

n performs TGC XMIT/REC ,DELAYS ,PULSING

n A/D OUTPUT

探头控制板 Beamformer Controller PCB, A24 2D有干扰可能为这块板坏。

n DETERMINGES PRF TIMING

n ARRAY TGC

n CHANNEL CONTROL AND SEQUENCING

n BEAMFORMER TIMING

MECHANICAL ACQUISITION

机械2D板 Mechanical 2D acquisition PCB 传送MEC 2D和M-MODE,把机械DOPPLER信号传给DOPPLER电路板

机械DOPPLER板 把DOPPLER信号传输给DOPPLER 处理板 A17

电机控制板 MOTOR CONTROLLER PCB

伺服机械电路板 EFT SERVO PCB → IVT/TRT CONTROL

心电隔离板 ECG ISOLATION PCB 与PHYSIO SENSORS相连,传送PHYSIO信号到M-MODE/PHYSIO处理板 A3

信号处理部 PROCESSOR GROUP

射频输入模块电路板 RF INPUT MODULE PCB A19

探头信号传输到该板,处理后DOPPLER和COLOR DOPPLER 经IMAGE BUS 到A21 混合DOPPLER PCB处理。

2D/M-MODE处理板 ECHO INPUT MODULE PCB A18

处理2D 和M-MODE 信号来自A19,传送黑白图象。、

混合DOPPLER处理板 HYBRID DOPPLER PCB A21

DOPPLER 处理板 DOPPLER PROCESSOR PCB 传送SPECTRAL频谱。A17

M-MODE /PHYSIO PROCESSOR PCB 处理和传送M-MODE /PHYSIO A3

WALL FILTER PCB 处理COLOR DOPPLER滤波 A16

频谱估算板 SPECTRAL ESTIMATOR PCB 处理2D COLOR / COLOR M-MODE A15

信号处理部PROCESSOR GROUP 有三种设计:PHASE 1 COLOR ,PHASE 2 COLOR ,DIGITAL PLUS

显示部 DISPLAY GROUP

2D黑白显示板 GRAY SC INTERFACE PCB ,2D黑白信号处理后送到AVP板上

OAG GRAY PCB 输出地址发生器黑白板 A10

OAG COLOR PCB 输出地址发生器彩色板 A14

DUAL BUFFER MEMORY PCB A12 ?

彩色SC接口板 COLOR SC INTERFACE PCB 传送COLOR DOPPLER (2D)到AVP。 A13

彩色图象板 COLOR GRAPHICS PCB 传送M-MODE COLOR DOPPLER 到AVP。 A7

静态图象电路板 STATIC GRAPHICS PCB 传送ANNOTATION TO AVP。 A5

FRAME GRABBER PCB 桢抓取板,传送FRAME GRAB DATA A9

SCROLLING GRAPHICS PCB 传送SPECTRAL / M-MODE / PHYSIO TO AVP。 A6

HARDCOPY INTERFACE PCB TRANSMIT SEPCTRAL / M-MODE / PHYSIO TO LINESCAN RECORDER A2

ADVANCED VIDEO PROCESSOR PCB AVP A8

REAR PANEL MODULE 后面板

PULSER POWER SUPPLY 给HIGH VOLTAGE CONTROL 提供电压给六个通道板

POWER SUPPLY

有三个电源,数字电源、脉冲电源、模拟电源 DIGITAL POWER SUPPLY,PULSER POWER SUPPLY,ANALOG POWER SUPPLY。

脉冲电源控制探头+/-80V

数字电源控制电路板+/-5V

●UM 4 plus

Control group 控制部

A15 系统控制板CPU SYSTEM CONTROLLER PCB

功能:接收用户命令来自控制组件及脚踏开关

控制系统电路板及周围设备

控制驱动器数据命令

提供所有的测量和分析功能

生成报告及驱动打印机工作

KEYBOARD CONTROLLER 键盘控制板

功能:把用户输入信息通过该板传送给系统控制板A15

DISPLAY GROUP 显示部

A14 显示板 DISPLAY CONTROLLER PCB

功能:产生所有的TV定时信号

产生混合视频

产生字符及图象

产生视频多路传输2D及M-MODE

图象增强的后处理

把图象信号给监视器,并且把A/V给辅助监视器及外围设备

控制实时图象和VCR回放的视频输出

A12~13 M模式和DOPPLER模式显示转换 TM-MODE 2D SCAN CONVERTER

A12 TM-MODE PCB 无此板,M模无法显示

存贮M-MODE AND DOPPLER

处理、格式M-MODE, DOPPLER ,ECG数据

传送M-MODE,DOPPLER,ECG 到周围设备

A13 TM-QUXILIARY PCB

产生注释及字符为滚动的M-MODE,DOPPLER

A9~A11 二维图象显示转换 2D SCAN CONVERTER

A9 S.C.RFACE,COMBO OR CINE PCB

读写功能记忆 memory read/write control

receives control,acquisition and address data from beamformer,pulse processor,and motor and system controllers

acquisition control of 2D and tm-mode scan converter groups via acq and sample clocks

frame averaging(smoothing)

interpolates sector coordinates to rectangular coordinates(row and column)for display

A10 扫描转换缓存板 SCAN CONVERTER BUFFER PCB

存储冻结桢2D图象及图象存储功能

stores 2D images for freeze frame and image store’function

2D scan conversion,shifts a line of video in at ultrasound scan rate(1-10khz),shifts out at TV horizontal sweep rate.

A11 输出地址发生器 OUTPUT ADDRESS GENERATOR

产生FRAME RATE

manipulates display addresses to produce pan/zoom magnification,image boundary limits.and other display features.

A5 脉冲处理板 Pulse processer pcb

S1. system receiver functions

Other functions same as pulse processor

S2. generates 2D XMT pulses for all scanheads

Controls power output

Generates TGC for sys rcvr and bmfmr front end pcb(A2)

Converts analog rf into digitized echo data

Edge enhancement/dynamic range

A7 Doppler processor PCB

Generate cw and pulsed Doppler signals from system receiver

Wall,thump and noise filtering

A/D conversion

Generates forward and reverse audio signals

A8 电机控制板 MOTOR CONTROLLER PCB

Positional control for mechanical scanheads

- end-fire(wobblers)

- access (rotary,incl dplx cx)and mercury

- annular array

sends xdcr fire to pulse processor pcb(A5)for all mechanical s/hs,incl AA

I.D’s mechanical scanheads

ACQUISITION GROUP 接收部

A1~A4 BEAMFORMER

A1 RF SELECT & DOPPLER PULSER PCB

与ANNULAR 探头、DFT、XDR ID相连

A2 Beamformer front end PCB

T/R Switchs block pulsers from reaching TGC

High coltage mux seitches (HV’s).Route pulsers to form correct aperture

TGC amplification of incoming rf signals

Sends eceived rf to beamformer pcb

A3 探头聚焦板

Crosspoint switches derotate (restores sequencial order of individual channel)

Removes delays from individual incoming rf signals

Depth of focus control

Demultipeses (combines)and amplifies rf signals

Sends linear rf to pulse processor

A4 探头控制板

Downloads operating parameters (depth,mode,s/h type,etc)from system controller to programmable pulse width generator

Controls beamforming

- phases aperture (creates parabolic beam,outer elements fire first;center elements fire last;via mechanical/analog delay lines)

- varies transmit focus by zone(varies delay ratios of parabolic beam curve)

- shifts aperture via hv mux switches on front end pcb

- shifts delays to maintain symmetry (rotation)

sends pulsers to beamformer front end pcb

controls beamformer focus pcb derotation (removing aperture phasing)

ID’s linear scanheads,incl AA

Sends liner xdr fire to pulse processor

键盘快捷键

Ctrl + E

用户默认菜单,可以调节各项设置

Ctrl + O

监视器左面显示去掉

Ctrl + S

检测录象带,ERROR! Diskette not inserted in Drive

Ctrl + L

出现版本号

Ctrl + ;

屏幕右面出现数行文字

Ctrl + Z

日期时间

Ctrl + U

是否保存探头参数


本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

ryw_win 发表于 2007-1-9 11:07 | 显示全部楼层
老哥,太老了点吧???呵呵!
莱茵石 发表于 2008-4-8 12:25 | 显示全部楼层
收藏
赵鹏 发表于 2008-4-8 14:07 | 显示全部楼层
老大,有人在灌水呢!
陈文辉 发表于 2008-5-27 13:41 | 显示全部楼层
强烈支持!顶~~
XSJQ 发表于 2015-5-12 22:52 | 显示全部楼层
很好,期待,学习中
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-27 01:58 , Processed in 0.936924 second(s), 33 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表