找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[求助]XS800i RBC采样错误、PLT采样错误

[复制链接]
huowawa 发表于 2012-6-6 18:35 | 显示全部楼层 |阅读模式

用户处一XS800i,09年6月份装机,现每日标本量200左右。4月底出现RBC采样错误报警,约每天报7、8次,当时清洗了RBC混匀室、检测部,对调了37-38与33-34电磁阀,仪器可以正常使用不报警了。近日,该仪器再次出现RBC采样错误、PLT采样错误,每天4、5次,重新对调了37-38与33-34阀,正常使用了2天后又开始报WBC采样错误。断断续续做过的处理:清洗1-2阀,清洗RBC混匀室、检测部,对调左侧2个1mL隔膜泵,现在还没有彻底解决。空气泵的状态:连续使用了近10W次,做测试时压力状态良好。以下贴报错标本数据,求帮忙分析:

以下为报警代码:

201,XS-800i^05347211^63184,2012/06/02,10:14:45,413020,-2147483648,-2147483648

1202,XS-800i^05347211^63184,2012/06/02,10:16:50,-413020,0,0

1203,XS-800i^05347211^63184,2012/06/02,10:41:38,413020,-2147483648,-2147483648

1204,XS-800i^05347211^63184,2012/06/02,10:43:41,413020,-2147483648,-2147483648

1205,XS-800i^05347211^63184,2012/06/02,10:43:41,413030,-2147483648,-2147483648

1206,XS-800i^05347211^63184,2012/06/02,10:45:55,-413020,0,0

1207,XS-800i^05347211^63184,2012/06/02,10:45:57,-413030,0,0

1208,XS-800i^05347211^63184,2012/06/02,11:31:50,413020,-2147483648,-2147483648

1209,XS-800i^05347211^63184,2012/06/02,11:31:50,413030,-2147483648,-2147483648

1210,XS-800i^05347211^63184,2012/06/02,11:32:55,-413020,0,0

1211,XS-800i^05347211^63184,2012/06/02,11:32:57,-413030,0,0

1212,XS-800i^05347211^63184,2012/06/02,11:58:50,413020,-2147483648,-2147483648

1213,XS-800i^05347211^63184,2012/06/02,11:58:50,413030,-2147483648,-2147483648

1214,XS-800i^05347211^63184,2012/06/02,11:59:18,-413020,0,0

1215,XS-800i^05347211^63184,2012/06/02,11:59:20,-413030,0,0

1216,XS-800i^05347211^63184,2012/06/02,15:24:48,413020,-2147483648,-2147483648

1217,XS-800i^05347211^63184,2012/06/02,15:24:51,-413020,0,0

1218,XS-800i^05347211^63184,2012/06/02,22:16:04,311240,151,-2147483648

1219,XS-800i^05347211^63184,2012/06/02,22:18:39,-311240,0,0

 

220,XS-800i^05347211^63184,2012/06/03,09:50:38,311240,107,-2147483648

1221,XS-800i^05347211^63184,2012/06/03,09:51:05,413020,-2147483648,-2147483648

1222,XS-800i^05347211^63184,2012/06/03,09:51:05,413030,-2147483648,-2147483648

1223,XS-800i^05347211^63184,2012/06/03,09:51:31,-311240,0,0

1224,XS-800i^05347211^63184,2012/06/03,09:52:28,-413020,0,0

1225,XS-800i^05347211^63184,2012/06/03,09:52:31,-413030,0,0

1226,XS-800i^05347211^63184,2012/06/03,10:50:57,413020,-2147483648,-2147483648

1227,XS-800i^05347211^63184,2012/06/03,10:51:08,-413020,0,0

1228,XS-800i^05347211^63184,2012/06/03,10:55:14,413020,-2147483648,-2147483648

1229,XS-800i^05347211^63184,2012/06/03,10:56:59,-413020,0,0

1230,XS-800i^05347211^63184,2012/06/03,11:15:18,413020,-2147483648,-2147483648

1231,XS-800i^05347211^63184,2012/06/03,11:17:32,413020,-2147483648,-2147483648

1232,XS-800i^05347211^63184,2012/06/03,11:17:32,413030,-2147483648,-2147483648

1233,XS-800i^05347211^63184,2012/06/03,11:23:58,-413020,0,0

1234,XS-800i^05347211^63184,2012/06/03,11:23:59,-413030,0,0

1235,XS-800i^05347211^63184,2012/06/03,11:48:12,413020,-2147483648,-2147483648

1236,XS-800i^05347211^63184,2012/06/03,11:48:27,-413020,0,0

1237,XS-800i^05347211^63184,2012/06/03,11:49:19,413020,-2147483648,-2147483648

1238,XS-800i^05347211^63184,2012/06/03,11:49:19,413030,-2147483648,-2147483648

1239,XS-800i^05347211^63184,2012/06/03,11:52:45,-413020,0,0

1240,XS-800i^05347211^63184,2012/06/03,11:52:47,-413030,0,0

1241,XS-800i^05347211^63184,2012/06/03,12:13:31,413020,-2147483648,-2147483648

1242,XS-800i^05347211^63184,2012/06/03,12:13:31,413030,-2147483648,-2147483648

1243,XS-800i^05347211^63184,2012/06/03,12:14:09,-413020,0,0

1244,XS-800i^05347211^63184,2012/06/03,12:14:11,-413030,0,0

1245,XS-800i^05347211^63184,2012/06/03,14:42:24,131029,-2147483648,-2147483648

1246,XS-800i^05347211^63184,2012/06/03,14:47:26,-131029,0,0

 

1285,XS-800i^05347211^63184,2012/06/06,09:07:12,311240,113,-2147483648

1286,XS-800i^05347211^63184,2012/06/06,09:07:13,-311240,0,0

1287,XS-800i^05347211^63184,2012/06/06,09:07:15,311240,127,-2147483648

1288,XS-800i^05347211^63184,2012/06/06,09:07:41,413020,-2147483648,-2147483648

1289,XS-800i^05347211^63184,2012/06/06,09:07:41,413030,-2147483648,-2147483648

1290,XS-800i^05347211^63184,2012/06/06,09:11:15,-413020,0,0

1291,XS-800i^05347211^63184,2012/06/06,09:11:16,-413030,0,0

1292,XS-800i^05347211^63184,2012/06/06,09:11:43,-311240,0,0

1293,XS-800i^05347211^63184,2012/06/06,09:32:07,413020,-2147483648,-2147483648

1294,XS-800i^05347211^63184,2012/06/06,09:32:26,-413020,0,0

1295,XS-800i^05347211^63184,2012/06/06,09:32:40,311240,113,-2147483648

1296,XS-800i^05347211^63184,2012/06/06,09:32:43,-311240,0,0

1297,XS-800i^05347211^63184,2012/06/06,09:33:08,413020,-2147483648,-2147483648

1298,XS-800i^05347211^63184,2012/06/06,09:33:08,413030,-2147483648,-2147483648

1299,XS-800i^05347211^63184,2012/06/06,09:33:17,-413020,0,0

1300,XS-800i^05347211^63184,2012/06/06,09:33:19,-413030,0,0

1301,XS-800i^05347211^63184,2012/06/06,09:37:22,413020,-2147483648,-2147483648

1302,XS-800i^05347211^63184,2012/06/06,09:37:25,-413020,0,0

1303,XS-800i^05347211^63184,2012/06/06,16:13:39,415000,-2147483648,-2147483648

1304,XS-800i^05347211^63184,2012/06/06,16:13:44,-415000,0,0

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

郑振寰 发表于 2012-6-6 21:13 | 显示全部楼层

这个修法真让人无奈,对换而且关键位置和关键位置对换,你能分析出什么来?既然37号阀换到34号阀,故障转移那就修阀或者换吧。

维修数据显示:数据逐渐降低,只能是鞘流或者样本注射出现问题,压力逐渐降低所致。至于是鞘流注射器还是稀释液罐的压力还是相关电磁阀不好导致的,就需要挨个检查了,XS系列和POCH系列最好不好换阀,直接修一下就行了,换来换去自己都不知道怎么回事了,反而把自己绕进去。

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
jackiehuang 发表于 2012-6-7 06:44 | 显示全部楼层
个人建议:换换23、24这两个阀,看看效果。
 楼主| huowawa 发表于 2012-6-7 08:21 | 显示全部楼层
不好意思,发帖太着急,关键部分打错了,“近日,该仪器再次出现RBC采样错误、PLT采样错误,每天4、5次,重新对调了37-38与33-34阀,正常使用了2天后又开始报WBC采样错误。"应该是“正常使用了2天后又开始报RBC采样错误”,因为没能看出故障转移的迹象,所以没有直接换掉电磁阀。
阿色 发表于 2012-6-7 09:10 | 显示全部楼层
把鞘流注射器丝杆和马达的螺杆上的灰尘去一去,还有那块铁片垂直移动的轨道也擦擦。把密封圈拆下来看看脏不脏,必要时换一新的。
wanylnajzm 发表于 2012-6-15 21:10 | 显示全部楼层
为啥要换23、24啊?高手指点。
徐峰-台州 发表于 2012-6-17 12:14 | 显示全部楼层
23还是24给忘了,是给稀释液瓶加压力的,也就是给红细胞、白细胞加鞘流的。
 楼主| huowawa 发表于 2012-6-18 22:53 | 显示全部楼层

今天去医院尝试将23-24阀与19-20阀对调,对调后立马报“废液瓶1错误”,从而确认23-24阀存在问题,只好更换掉了。

谢谢各位指导!

earl007 发表于 2015-5-28 10:05 | 显示全部楼层
哎呀看的好晕
Fensun 发表于 2023-7-29 00:48 | 显示全部楼层
学习了
fafafafafafa 发表于 2023-8-12 17:00 | 显示全部楼层
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-4-1 20:49 , Processed in 0.660757 second(s), 39 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表