找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

SYSMEX XT1800I/2000I FLAG 提示信息

[复制链接]
郑振寰 发表于 2005-3-30 00:10 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×
<XE-2100>
  Flagging Item Judgement item for Flagging
WBC WBC Abn Scattergram  
   ①Each cluster cannot be differentiated on DIFF scattergram. DIFFch
   ②Each cluster cannot be differentiated on BASO scattergram. BASOch
   ③5DIFF data cannot be calculated. 3BC#,4DIFF#,4DIFF%
   ④WBC-related data cannot be corrected by NRBC. LYMPH#,NRBC#
  NRBC Abn Scatteagram NRBCch
  Neutropenia NEUT#,NEUT%
  Neutrophilia NEUT#,NEUT%
  Lymphopenia LYMPH#,LYMPH%
  Lymphocytosis LYMPH#,LYMPH%
  Monocytosis MONO#,MONO%
  Eosinophilia EO#,EO%
  Basophilia BASO#,BASO%
  Leukocytopenia WBC#
  Leukocytosis WBC#
  NRBC present NRBC%
  Blasts? DIFF,IMIch,WBC#
  Immature Gran? DIFF,IMIch,WBC#
  Left Shift? DIFF,IMIch,WBC#
  Atypical Lympho? DIFF,IMIch,WBC#
  Abn Lympho/ L-Blasts? DIFF,IMIch,WBC#
  NRBC? DIFFch
  RBC Lyse resistance? DIFF,WBC#
RBC RBC Abn Distribution RBCch,WBC#
  Dimorphic Population RBCch
  RET Abn Scattergram RETch
  Reticulocytosis RET#,RET%
  Anisocytosis RDW-SD,RDW-CV
  Microcytosis MCV
  Macrocytosis MCV
  Hypochromia MCHC
  Anemia HGB
  Erythrocytosis RBC#
  RBC Aggulutination? MCHC,MCH,RBC#,RU(%)
  Turbudity/HGB Interference? MCHC
  Iron Deficiency? MCHC,MCV,RDW-CV
  HGB Defect? MCV,RDW-CV
  Fragments?
PLT PLT Abn Scattergram RETch
  PLT Abn Distribution PL(%),PU(%),PDW-SD,PDW,PLCR,MPV,PU
  Thrombocytopenia PLT#
  Thrombocytosis PLT#
  PLT Clumps? DIFF,IMIch
  PLT Clumps(S)? PLT#,PDW,PL(%),PU(%),P-LCR,MPV,PU
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2005-3-30 00:11 | 显示全部楼层
<XT-2000i>
  Flagging Item Judgement Item for Flagging
WBC WBC Abn Scattergram  
   ①Each cluster cannot be differentiated on DIFF scattergram. DIFFch
   ②Each cluster cannot be differentiated on BASO scattergram. BASOch
   ③5DIFF data cannot be calculated. 3BC#,4DIFF#,4DIFF%
   ④WBC-related data cannot be corrected by NRBCs LYMPH#
  NRBC Abn Scatteagram  
  Neutropenia NEUT#,NEUT%
  Neutrophilia NEUT#,NEUT%
  Lymphopenia LYMPH#,LYMPH%
  Lymphocytosis LYMPH#,LYMPH%
  Monocytosis MONO#,MONO%
  Eosinophilia EO#,EO%
  Basophilia BASO#,BASO%
  Leukocytopenia WBC#
  Leukocytosis WBC#
  NRBC present  
  Blasts? DIFFch
  Immature Gran? DIFFch
  Left Shift? DIFFch
  Atypical Lympho? DIFFch
  Abn Lympho/ L-Blasts? DIFFch
  NRBC? DIFFch
  RBC Lyse resistance? DIFF,WBC#
RBC RBC Abn Distribution RBCch,WBC#
  Dimorphic Population RBCch
  RET Abn Scattergram RETch
  Reticulocytosis RET#,RET%
  Anisocytosis RDW-SD,RDW-CV
  Microcytosis MCV
  Macrocytosis MCV
  Hypochromia MCHC
  Anemia HGB
  Erythrocytosis RBC#
  RBC Aggulutination? MCHC,MCH,RBC#,RU(%)
  Turbudity/HGB Interference? MCHC
  Iron Deficiency? MCHC,MCV,RDW-CV
  HGB Defect? MCV,RDW-CV
  Fragments?
PLT PLT Abn Scattergram RETch
  PLT Abn Distribution PL(%),PU(%),PDW-SD,PDW,PLCR,MPV,PU
  Thrombocytopenia PLT#
  Thrombocytosis PLT#
  PLT Clumps? DIFFch
  PLT Clumps(S)? PLT#,PDW,PL(%),PU(%),P-LCR,MPV,PU
  The flag which is available only when with NRBC/IMIch.
  The flag which is affected in Flagging rule when without NRBC/IMIch
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-3-30 00:12 | 显示全部楼层
The Flags of NRBC/IMIch
  Flagging Item Judgement Conditions Related Ch
1 WBC Abn Scattergram A A LYMPH<NRBC DIFF,NRBC
   ④WBC-related data cannot be corrected by NRBC        
2 NRBC Abn Scatteagram (A and B) or C A the peak height of WBC population≧100 NRBC
    B the height% of discrimination between Gst and WBC≧50 NRBC
      C the number of particle in N_Ghost area≧50 NRBC
3 NRBC present A A NRBC%>2% NRBC
4 Blasts? (A or B) A IMI-Blasts? IMI
  and(C or D) B IMI-I.G.? IMI
    C dBl#≧80 DIFF
      D dBl%≧2% DIFF
5 Immature Gran? A or B or C A dIG#≧60 DIFF
    B dIG%≧3% DIFF
      C IMI-I.G.? IMI
6 Left Shift? A or B or C A dLS#≧200 DIFF
    B dLS%≧4% DIFF
      C I_Area1≧400 IMI
7 Atypical Lympho? (A and B) and A not IMI-Blasts? IMI
  (C or D or E) B not IMI-I.G.? IMI
    C dAl#≧100 DIFF
    D dAl%≧1% DIFF
      E dAl2#≧25 DIFF
8 Abn Lympho/ L-Blasts? (A and B) and C A not IMI-Blasts? IMI
    B not IMI-I.G.? IMI
      C the number of particle on Lym:Mon border≧20 DIFF
9 PLT Clumps? A or B or C or A dPc#≧200 DIFF
  D or E or F B dPc%≧3%@DIFF_WBC≧500 DIFF
    C the number of particle on Gst/Baso border≧15 DIFF
    D iPLTC#≧100 IMI
    E iPLTCa#≧10 IMI
      F iPLTCc#+iPLTCd#≧10 IMI
  The flag which is available only when with NRBC/IMIch.
  The flag which is affected in Flagging rule when without NRBC/IMIch
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-3-30 00:14 | 显示全部楼层

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
 楼主| 郑振寰 发表于 2005-3-30 00:16 | 显示全部楼层

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
 楼主| 郑振寰 发表于 2005-3-30 00:17 | 显示全部楼层

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
 楼主| 郑振寰 发表于 2005-3-30 00:19 | 显示全部楼层

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-4-1 17:03 , Processed in 0.576462 second(s), 32 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表