找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

cx系列加热器板说明

[复制链接]
郑振寰 发表于 2005-10-30 21:58 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

Heater Boards (443271 and 443272) with TI Chips

Chapter: Synchron CX Documents

Summary

1/21/98

SUPERSEDES PRODUCT

NONE CX3/3D/7/7D/9

CLASSIFICATION SERVICE DOC - MANDATORY

S/N AFFECTED CX3 Delta:1930 to 1962

CX3 RIP: 874, 1080, and 1620

CX3 Recon:2774 and 4493

CX7 RIP: 481, 2851, 3308, and 3465

CX7 Recon:3471, 3863, 4438, 4663, and 4820

CX9 ALX: 7729 to 7770

P/N(S)/ 443271/ CREATININE/CAL HEATER CONTROLLER

DESCRIPTION(S) 443272/ BUN/GLUCOSE HEATER CONTROLLER

ESTIMATED LABOR HOURS 1 Hour

SPECIAL NOTES:

For all Instruments listed above and all heater boards

ordered between December 5, 1997 and January 8, 1998, the

four heater boards on the cup chemistries need to be

visually checked. If the board contains a Texas Instrument

IC chip, it must be replaced.

After completing the replacement of boards or the inspection

with no replacement boards needed, the Engineer is to Email

Terry Edson at tledson@beckman.com with the following

information:

Facility name and country.

A. Instrument Serial Number.

B. Number of boards replaced (if any).

C. Date boards replaced/inspection.

SYMPTOMS:

Slow temperature recovery. This could produce a WRONG

ANSWER with this board.

Manufacturing has found a non-validated IC chip on the

heater board assemblies. These IC s can be identified by

the TI symbol on the chip (looks like the state of Texas

with an I on it).

Any heater board manufactured between December 5, 1997 and

January 8, 1998 could have this chip and should be checked.

All inventories have been reworked to remove the bad chips.

Approximately 25 replacement heater boards have been sent

out that could have this Bad chip.

WARNING:

USE STATIC PRECAUTIONS WHEN WORKING WITH ALL

BOARDS. THIS INCLUDES WEARING AN ESD WRIST

STRAP AND WORKING ON A STATIC FREE MAT.

Engineers who provide primary support for the instrument

with serial numbers listed are to contact:

Hank Roberts Phone (714) 961-3847

EMAIL: HMROBERTS@BECKMAN.COM

Quality Control

RE: TI heater board replacement

(Please specify instrument serial number and number of

boards needed when calling)

2. After completing the replacement of boards or the

inspection with no replacement boards needed, the Engineer

is to Email Terry Edson at tledson@beckman.com with the

following information:

Facility name and country.

A. Instrument Serial Number.

B. Number of boards replaced (if any).

C. Date boards replaced/inspection.

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-1 06:20 , Processed in 0.771835 second(s), 32 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表