找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转帖]麻醉机蒸发器知识--英文

[复制链接]
郑振寰 发表于 2010-3-2 15:57 | 显示全部楼层 |阅读模式

Anaesthetic vaporiser

 

 
Anaesthetic vaporiser

The design of these devices takes account of varying

  • ambient temperature
  • fresh gas flow
  • agent vapour pressure

Contents

  • 1 Historical vaporisers
  • 2 Modern vaporisers
  • 3 Plenum vaporisers
  • 4 Drawover vaporisers
  • 5 Dual-circuit gas-vapour blender

Historical vaporisers

Historically, ether (the first volatile agent) was first used by John Snow's inhaler (1847) but was superseded by the use of chloroform (1848). Ether then slowly made a revival (1862–1872) with regular use via Curt Schimmelbusch's "mask", a narcosis mask for dripping liquid ether. Now obsolete, it was a mask constructed of wire, and covered with cloth.

Pressure and demand from dental surgeons for a more reliable method of administrating ether helped modernise its delivery. In 1877, Clover invented an ether inhaler with a water jacket, and by the late 1899 alternatives to ether came to the fore, mainly due to the introduction of spinal anaesthesia. Subsequently this resulted in the decline of ether (1930–1956) use due to the introduction of cyclopropane, trichloroethylene, and halothane. By the 1980s, the anaesthetic vaporiser had evolved considerably; subsequent modifications lead to a raft of additional safety features such as temperature compensation, a bimetallic strip, temperature-adjusted splitting ratio and anti-spill measures.

 Modern vaporisers

There are generally two types of vaporisers: plenum and drawover. Both have distinct advantages and disadvantages. A third type of vaporiser is exclusively used for the agent desflurane.

Plenum vaporisers

The plenum vaporiser is driven by positive pressure from the anaesthetic machine, and is usually mounted on the machine. The performance of the vaporiser does not change regardless of whether the patient is breathing spontaneously or is mechanically ventilated. The internal resistance of the vaporiser is usually high, but because the supply pressure is constant the vaporiser can be accurately calibrated to deliver a precise concentration of volatile anaesthetic vapour over a wide range of fresh gas flows. The plenum vaporiser is an elegant device which works reliably, without external power, for many hundreds of hours of continuous use, and requires very little maintenance.

The plenum vaporiser works by accurately splitting the incoming gas into two streams. One of these streams passes straight through the vaporiser in the bypass channel. The other is diverted into the vaporising chamber. Gas in the vaporising chamber becomes fully saturated with volatile anaesthetic vapour. This gas is then mixed with the gas in the bypass channel before leaving the vaporiser.

A typical volatile agent, isoflurane, has a saturated vapour pressure of 32kPa (about 1/3 of an atmosphere). This means that the gas mixture leaving the vaporising chamber has a partial pressure of isoflurane of 32kPa. At sea-level (atmospheric pressure is about 101kPa), this equates conveniently to a concentration of 32%. However, the output of the vaporiser is typically set at 1-2%, which means that only a very small proportion of the fresh gas needs to be diverted through the vaporising chamber (this proportion is known as the splitting ratio). It can also be seen that a plenum vaporiser can only work one way round: if it is connected in reverse, much larger volumes of gas enter the vaporising chamber, and therefore potentially toxic or lethal concentrations of vapour may be delivered. (Technically, although the dial of the vaporiser is calibrated in volume percent (e.g. 2%), what it actually delivers is a partial pressure of anaesthetic agent (e.g. 2kPa)).

The performance of the plenum vaporiser depends extensively on the saturated vapour pressure of the volatile agent. This is unique to each agent, so it follows that each agent must only be used in its own specific vaporiser. Several safety systems, such as the Fraser-Sweatman system, have been devised so that filling a plenum vaporiser with the wrong agent is extremely difficult. A mixture of two agents in a vaporiser could result in unpredictable performance from the vaporiser.

Saturated vapour pressure for any one agent varies with temperature, and plenum vaporisers are designed to operate within a specific temperature range. They have several features designed to compensate for temperature changes (especially cooling by evaporation). They often have a metal jacket weighing about 5kg, which equilibrates with the temperature in the room and provides a source of heat. In addition, the entrance to the vaporising chamber is controlled by a bimetallic strip, which admits more gas to the chamber as it cools, to compensate for the loss of efficiency of evaporation.

The first temperature-compensated plenum vaporiser was the Cyprane 'FluoTEC' Halothane vaporiser, released onto the market shortly after Halothane was introduced into clinical practice in 1956.

Drawover vaporisers

The drawover vaporiser is driven by negative pressure developed by the patient, and must therefore have a low resistance to gas flow. Its performance depends on the minute volume of the patient: its output drops with increasing minute ventilation.

The design of the drawover vaporiser is much simpler: in general it is a simple glass reservoir mounted in the breathing attachment. Drawover vaporisers may be used with any liquid volatile agent (including older agents such as diethyl ether or chloroform, although it would be dangerous to use desflurane). Because the performance of the vaporiser is so variable, accurate calibration is impossible. However, many designs have a lever which adjusts the amount of fresh gas which enters the vaporising chamber.

The drawover vaporiser may be mounted either way round, and may be used in circuits where re-breathing takes place, or inside the circle breathing attachment.

Drawover vaporisers typically have no temperature compensating features. With prolonged use, the liquid agent may cool to the point where condensation and even frost may form on the outside of the reservoir. This cooling impairs the efficiency of the vaporiser. One way of minimising this effect is to place the vaporiser in a bowl of water.

The relative inefficiency of the drawover vaporiser contributes to its safety. A more efficient design would produce too much anaesthetic vapour. The output concentration from a drawover vaporiser may greatly exceed that produced by a plenum vaporiser, especially at low flows. For safest use, the concentration of anaesthetic vapour in the breathing attachment should be continuously monitored.

Despite its drawbacks, the drawover vaporiser is cheap to manufacture and easy to use. In addition, its portable design means that it can be used in the field or in veterinary anaesthesia.

Dual-circuit gas-vapour blender

The third category of vaporiser (a dual-circuit gas-vapour blender) was created specifically for the agent desflurane. Desflurane boils at 23.5C, which is very close to room temperature. This means that at normal operating temperatures, the saturated vapour pressure of desflurane changes greatly with only small fluctuations in temperature. This means that the features of a normal plenum vaporiser are not sufficient to ensure an accurate concentration of desflurane. Additionally, on a very warm day, all the desflurane would boil, and very high (potentially lethal) concentrations of desflurane might reach the patient.

A desflurane vaporiser (e.g. the TEC 6 produced by Datex-Ohmeda) is heated to 39C and pressurised to 200kPa (and therefore requires electrical power). It is mounted on the anaesthetic machine in the same way as a plenum vaporiser, but its function is quite different. It evaporates a chamber containing desflurane using heat, and injects small amounts of pure desflurane vapour into the fresh gas flow. A transducer senses the fresh gas flow.

A warm-up period is required after switching on. The desflurane vaporiser will fail if mains power is lost. Alarms sound if the vaporiser is nearly empty. An electronic display indicates the level of desflurane in the vaporiser.

The expense and complexity of the desflurane vaporiser have contributed to the relative lack of popularity of desflurane, although in recent years it is gaining in popularity.


本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

ganzheng 发表于 2020-7-11 11:11 | 显示全部楼层
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-18 08:10 , Processed in 0.289045 second(s), 33 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表