找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

Elecsys 2010与其它品牌比较

[复制链接]
郑振寰 发表于 2005-1-10 11:59 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×
Competitor Specification List
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
system size 130 x 73 x 56 160 x 85 x 130 cm 139 x 66 cm 61 x 152 x 59 cm 106 x 63 x 40 cm
technology electro chemiluminescence FPIA, MEIA, ICIA alk. phosphatase + Lumiphos acridinium chemiluminescence alk.phosphatase + dioxethane
random access yes yes yes yes yes
continuous access yes yes yes yes yes
operation mode random access multibatch multi-batch batch batch, multibatch
LLD sample, reagent, system fluids sample, reagent ? sample, reagent sample, reagent
temperature control incubator, reagent, detection unit, system fluids reagent reagent; incubation incub., reagent, (sample, detec.) incub., reagent
waste tips immense amount of solid waste low waste production waste container too small ?
bound/free seperation magnetic separation within measuring cell glass fiber matrix cell magnetic separation in 3 steps magnetic separation in cuvettes spinning of reaction vessel
reagents ready-to-use yes yes yes yes yes
  temp. equilibr. required yes no yes no yes
  on-board stability up to 8 weeks 112 hrs continously 28 d 1.7 d 30 d
  shelf life 6 - 12 months 90 days 365 d 270 d 30 - 180 d
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2005-1-10 12:00 | 显示全部楼层
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
  barcode yes yes yes yes yes
  monitoring of remaining tests yes yes yes by calculation yes by physical means yes by physical means
calibration ready-to-use yes yes yes no some
  stability 12 weeks 28 d 14 - 28 d 1 - 28 d 1 - 4 weeks
  warning before expiry yes no no no yes
  calibrators required per analyte 2 1 - 2 / 4 - 12 1 - 7 1 - 2 1 - 2
  time required for calibration 10 - 20 min 15 - 35 min 8 - 10 min 15 min 42 - 72 min
  calibration 2 point cal. cal. concept unclear 6 point 2 point recal. with multical. 2 point recal. in quadruplicate
  recalibration no 2 point no 2 point (every 7 d) 2 point (every 30 d)
  full calibration never never yes (6 -7 cal.; every 21 - 30 d) no no
  lot calibration yes yes ? ? ?
  master calibration curve positive master curve identification seperate package insert (extern. read) not available seperate package insert (extern. read) not available
  multicalibrator no no ? identicat. field specific yes
samples STAT yes 60 - 90 samples 60 60 samples 35 samples
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-1-10 12:00 | 显示全部楼层
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
             
  TAT 9 - 18 min 23 - 45 min 20 - 45 min 20 - 40 min 44 - 70 min
  continous loading yes yes yes no yes
  loading capacity 30 (75: rack) 90 60 samples 60 47
  sample handling 30 prim.tubes (rotor)/15 racks 6 segments each for 10 prim. or 15 sec. 6 segments for 10 primary tubes sample rotor 60 posit. for prim. continuous chain for 85 cups + test units
  primary cups yes yes yes yes no
  PSID internal barcode reader internal barcode reader internal (if host connection)/external wand internal barcode reader no PSID - barcoded cup-number
  PRID internal barcode reader for PDF 417 internal barcode reader external barcode wand internal barcode reader internal barcode reader
  parameters 18 31 25 25 32
  automatic predilution yes 1:10; 1:200 no yes no
  auto rerun when exceeding limits no yes no yes no
  rerun requested no yes yes yes no
  incubation time 4.5 - 9 min < = 20 min 5 - 20 min 7.5 - 15 min (2.5 - 7.5 min) 30 - 60 min
  incubation steps 2 2 - 3 2 - 3 1 - 2 2 - ?
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-1-10 12:01 | 显示全部楼层
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
  throughput theor. (pract.) 86/hr (70 - 80/hr) up to 120/hr (70 - 90/hr) 100/hr (50-60/hr) 180/hr (80 - 120/hr) 150/HR (<<100/hr)
  throughput limited bec. measuring cycle takes 42 sec. multi-batch / different ICT / STAT multibatch; different ICT/STAT; long turn around time extra rinse, many stops, STAT first result after 40 or 70 min
  time of first result 10 - 20 min 8 - 30 min 15 - 75 min 15 min 42 - 72
  results after 1 hr 57 - 72 60 - 100 0 - 75 100 - 130 0 - 45
  carry over no 2e-7, low bec. of needle washing 7.00E+07 none < 2e5
  sample volume 10 - 50 μ 30 - 200 μ 6 - 100 μ 10 - 200μ 5 - 75 μ
  dead volume 50 - 100 μl depending on cu 100 - 120 μ 100 μ 50 μ 100 μ
  assay mode quasi-homogenous homogenous/heterogenous heterogenous heterogenous heterogenous
  assay principle ECLIA FIA, FPIA LIA LIA LIA
software/data handling multitasking yes yes no yes yes
  host bidirectional bidirectional yes bidirectional bidirectional
  data input barcode/keyboard or host keyboard, barcode keyboard, barcode keyboard/barcode keyboard
  report patient-/testoriented patient-/testoriented patient-/testoriented patient-/testoriented patient oriented
  patient's statistics not yet yes yes yes yes
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-1-10 12:02 | 显示全部楼层
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
  QC avail. in 4/97 yes yes yes yes
  saving results 400 poss. on disk < = 1000 sample and 2200 assay results on disk ? ?
integration   rack conveyor (2 * E2010)        
    LSM ( ES, clin.chem.        
    aliquoter, sample distributor)        
maintenance hands-on-time daily 3 min 5 (min) 5 min 2 min < 5 min
  weekly 20 - 25 min 20 min 7 min 20 min 5 min
  monthly 25 min 10 hrs 15 min 30 min 15 min
  total time daily   15 min 15 min 48 min < 5 min
  weekly   40 min 45 min 20 min 5 min
  monthly   30 hrs 15 min 30 min 15 min
  daily start-up time   5 min 0 min 4 min < 5 min
    Roche Elecsys 2010 Abbott AxSYM Sanofi Diagnostics Pasteur ACCESS Ciba Corning ACS:180 Diagnostics Product Corporation Immulite
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
chencheng7 发表于 2005-10-28 21:41 | 显示全部楼层
楼主,能否搞成中文.俺也好更了解了解,谢啦!
朱靖熙 发表于 2008-4-4 11:02 | 显示全部楼层

还好,都能看懂啊。

不晓得 发表于 2008-7-20 11:45 | 显示全部楼层
大哥,看不懂E文。
復華 发表于 2008-8-4 11:21 | 显示全部楼层
現在正好在評估,缺資料,看到像挖到寶,感謝樓主
冯家湖 发表于 2008-8-11 00:27 | 显示全部楼层

15年前的资料

yeeb 发表于 2021-5-25 14:51 | 显示全部楼层

这就是中国医疗器械的差异
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-10-23 18:23 , Processed in 0.388585 second(s), 38 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表