3M 969/XP535 Errors - GE Exclusive Use Only GE Exclusive Use Only
      This material is the property of GE Medical Systems (GEMS) and contains
      information, which is restricted to use by GEMS' Manufacturing,
      Engineering and Service personnel only. Any reproduction in whole or in
      part is strictly prohibited. Neither this material nor any information
      contained hereinshall be disclosed to anyone outside GEMS Manufacturing,
      Engineering and Service personnel

952 & 959 Error Codes
Introduction: This document is intended as a quick reference only. Please use 3M
service manuals for troubleshooting. This document was formatted for display on
a Palm Pilot using iSilo. For any changes, additions or errors, please contact
Bob Flye. Last Revised 12/25/1999.
Error Select List

 

969/XP535 Error Select list
  969 Errors L39 - P164
  969 Errors L165 - K195
  XP535 Errors P501 - P633

 

969 Error Codes L39-L164/P164
      39 48 64 66 67 68 75 85 112 113
      114 115 116 117 118 119 121 122
      123 125 130 132 134 139 140 141
      151 152 154 155 161 162 163 164
Error Select List

 

 

969 Error Codes L165/P165-K195
      165 166 167 171 172 174 175 176
      177 179 200 201 202 210 211 212
      214 215 216 501 502 601 603 604
      605 606 608 609 620 622 623 910
      912 913 K1 K3 K4 K9 K58 K183
      K184 K192 K195
Error Select List

 

 

XP535 Error Codes P501-P633
      501 502 504 505 508 510 511 523
      524 525 526 529 530 531 532 533
      534 535 536 539 549 550 567 570
      571 572 589 590 624 630 631 632
      633
Error Select List

 

 


969 error code L39
EIB not detected.
Check power supply on EIB, fiber optic links. Try restoring from backup disk.
Swap with another EIB.

969 error code L48
EOF Timeout.
Check output module first, If you have a slave printer, switch output modules.
Could be cable to MCS board or MCS board.

969 error code L64
Acquisition Parity EIB.
Try disabling parity on input module. Could be EIB, fiber cables or input
module.

969 error code L66
Acquisition Out of Bounds.
Check RAM size jumpers, if OK, may be input module.

969 error code L67
FIFO Overflow.
See t/s flowchart.

969 error code L68
Timeout EIB.
Could be bad video, bad EIB or input module. See t/s guide.

969 error code L75
Local Panel Error.
Lack of 12 volts on local lanel. Check power supplies and wire harness. See t/s
guide.

969 error code L85
Memory Full.
Is camera able to output films? Check input module for bad blocks.

969 error code L112/P112
Rewind Unit Error.
Supply cartridge does not open. Check power supplies, check for mechanical
binding. Consult t/s guide.

969 error code L113/P113
Rewind Unit Trig.
Problem with opening supply magazine. See t/s guide.

969 error code L114/P114
Rewind Unit 8 Error.
Problem with opening supply magazine. See t/s guide.

969 error code L115/P115
Pickup Cup Arm Lower Error.
Bottom sensor not actuated. Check supplies, sensor and wire harness. See t/s
guide.

969 error code L116/P116
Pickup Cup Arm Upper Error.
Check top sensor adjustment, check for sheared screw on pickup arm motor gear,
see t/s guide.

969 error code L117/P117
Pickup Cup Slide Error.
Outer cups not moving. Check supplies, pickup arm harness, sensor or PWA-B. See
t/s guide.

969 error code L118/P118
Width Slide Guide Error.
Check supplies, pickup arm harness or PWA-B.

969 error code L119/P119
Pickup Arm Error.
Problem picking up the film and getting it into the rollers. See t/s guide.

969 error code L121/P121
Attenuator Error.
Problem between optics module and cpu. See t/s guide.

969 error code L122/P122
No Optics Module.
See t/s guide to isolate between optics module and PWA-A or PWA-B.

969 error code L123/P123
Polygon Motor Error.
Check MCS supplies, and polygon signals to isolate problem.

969 error code L125/P125
No SOS, (Start of Scan)
Check sos pulses at optics module PWA-S, see t/s guide.

969 error code P130
Magazine Door Open.
See t/s guide to isolate supply drawer error.

969 error code P132
No Magazine.
Camera does not sense supply cartridge is installed, see t/s guide.

969 error code P134
Magazine Empty.
See t/s guide. Problem may be that the NVRAM thinks the magazine has been used.
To clear barcode NVRAM: On MCS board, set sw5 to B and sw4 to 0, press sw2.

969 error code L139/P139
Bad Barcode.
See t/s guide.

969 error code P140
Receive Magazine Door Open.
Happens lots, have customer open receive door and close again to open magazine.
They may have to do it a few times. On site, check t/s guide.

969 error code P141
No Receive Magazine.
See t/s guide.

969 error code L151/P151
A/D Conversion Error.
Check MCS power supplies, replace MCS.

969 error code L152/P152
Linebuff1 Error.
Check MCS power supplies, replace PWA-A.

969 error code L154/P154
NVRAM Error.
Reset PWA-A to defaults, sw4 to 0, sw5 to B then press reset. Set sw4/5 back to
0. If still bad, replace PWA-A.

969 error code L155/P155
Power Failure.
Check MCS power supplies, replace MCS.

969 error code L161/P161
Jam Magazine.
See note in t/s guide, if L161 is in log, do L119/P119 checks also.

969 error code L162/P162
Jam Transport.
Film jam. See t/s guide.

969 error code L163/P163
Jam To Exposure.
Jam 2 sensor detects jam. See t/s guide.

969 error code L164/P164
Jam Exposure.
Leading/trailing edge amp problem. See t/s guide.

969 error code L165/P165
Jam to Processor.
Jam with film going into processor, check jam 3 sensor. See t/s guide.

969 error code L166/P166
Pickup Tries.
Possible problem with pickup mid sensor, harness or power supplies. See t/s
guide.

969 error code L167
Double Feed.
Read note in t/s guide. The guide does not address film, always check film.

969 error code L171/P171
Supply Drawer Open Error.
See t/s guide.

969 error code L172/P172
Receive Magazine Door Open Error.
See t/s guide.

969 error code L174/P174
Receive Magazine Close Error.
Possible bad switch or harness.

969 error code L175/P175
Receive Magazine Open Error.
Lots going on, see t/s guide.

969 error code L176/P176
Supply Cover Open Error.
Lots going on, see t/s guide.

969 error code L177/P177
Supply Cover Close Error.
Check for obstruction or binding, see t/s guide.

969 error code P179
Jam Access Open.
Access panel interlock switch problem, see t/s guide.

969 error code L200
Parity Error.
Probable input module, possible output module. See t/s guide.

969 error code L201
Sync Error.
Probable input module, possible output module. See t/s guide.

969 error code L202
Print Out of Bounds.
Probable input module, possible output module. See t/s guide.

969 error code L210
Buffer Overflow.
Probable input module, possible output module. See t/s guide.

969 error code L211
EOL Error.
Probable input module, possible output module or PWA-A. See t/s guide.

969 error code L212
Handshake Error.
Probable input module, possible output module. See t/s guide.

969 error code L214
Line Length Error.
Probable input module, possible output module. See t/s guide.

969 error code L215
Output Error.
Probable input module, possible output module or PWA-A. See t/s guide.

969 error code L216
Realtime Error.
Input module or output module problem. See t/s guide.

969/XP535 error code L501/P501
Dock Cover Open.
Interlock problem, Docking pins not fully engaged. See t/s guide.

969/XP535 error code P502
No Communication With Processor.
Output module problem or connection to processor. Check 9 pin plug between
imager and processor at bottom front. Open processor front door to locate. See
Also P590 in the XP535 Service manual.

XP535 error code P504
Developer Replenishment Tank Low.
Bad sensor switch or Processor Control PWA. See t/s flowchart.

XP535 error code P505
Fixer Replenishment Tank Low.
Bad sensor switch or Processor Control PWA. See t/s flowchart.

XP535 error code P508
Excessive Water Flow.
Check level sensor or control electronics. See t/s flowchart.

XP535 error code P510
System Calibration.
Normal status message.

XP535 error code P511
Developer or Fixer Bath Warming/Cooling Error.
See t/s flowchart, use MPC Diag to isolate problem.

XP535 error code P523
No Developer Heat.
Probable processor Control PWA or temp probe. See t/s flowchart.

XP535 error code P524
Developer Heat Stuck On.
Probable processor Control PWA or temp probe. See t/s flowchart.

XP535 error code P525
Low Developer Level.
Check level sensor, if OK then see t/s flowchart.

XP535 error code P526
Developer Replenishment Tank Empty.
If tank is OK, replace Processor Control PWA. See t/s flowchart.

XP535 error code P529
Developer Temperature Error.
Probable processor Control PWA or temp probe. See t/s flowchart.

XP535 error code P530
Dryer Heat Inoperative.
See t/s flowchart.

XP535 error code P531
Dryer Heat Stuck On.
Probable cause Power Switching PWA or Processor Control PWA. See t/s flowchart.

XP535 error code P532
Dryer Temperature Error.
Check temp probe, could be Processor Control PWA or Power Switching PWA. See t/s
flowchart.

XP535 error code P533
No Fixer Heat.
See t/s flowchart.

XP535 error code P534
Fixer Heat Stuck On.
If probe is OK, check Power Switching PWA or Processor Control PWA. See t/s
flowchart.

XP535 error code P535
Low Fixer Level.
See t/s flowchart.

XP535 error code P536
Fixer Replensihment Tank Empty.
If tank is OK, replace Processor Control PWA. See t/s flowchart.

XP535 error code P539
Fixer Temperature Error.
If probe is OK, check Power Switching PWA or Processor Control PWA. See t/s
flowchart.

XP535 error code P549
Entry Sensor Malfunction.
Replace entry sensor.

XP535 error code P550
Exit Sensor Malfunction.
Replace exit sensor.

XP535 error code P567
Film Entering Processor.
Unexpected film is entering processor, status only.

XP535 error code P570
Transport Motor Malfunction.
See t/s flowchart.

XP535 error code P571
Transport Motor Slow.
See t/s flowchart.

XP535 error code P572
Low Water Flow.
Check level sensor, or could be Processor Control PWA or Power Switching PWA.
See t/s flowchart.

XP535 error code P589
Non-Volatile Ram Check.
Open and close cover, if error doesn't go away, replace Processor Control PWA.

XP535 error code P590
Processor Power Down.
See Explanation in service manual.

969 error code L601/P601
Invalid Power Monitor Offset.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code L602/P602
Invalid Power Monitor Range.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code L603/P603
Invalid Opt Density Range.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code L604/P604
Invalid Laser Dyn Range.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code L605/P605
Atten Test Fail.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code P606
Opt Command Fail.
Write default optics parameters to output module. If that fails, probable optics
module failure or PWA-A.

969 error code P608
Opt Unk Resp.
Run comm tests on output module. Probable output module or PWA-A.

969 error code P609
Opt Inv Data.
Run comm tests on output module. Probable output module or PWA-A.

969 error code L620/P620
No Film Model.
See t/s guide.

969 error code P622
Bad Table EPROM
Replace transfer function EPROM's or input module.

969 error code L623/P623
Bad Transfer Function.
Replace transfer function EPROM's or input module.

XP535 error code P624
Automatic Image Quality Control Malfuction.
Check densitometer, use denstest.exe to test. OLC should have a copy they can
put on globe. See t/s flowchart.

XP535 error code P630
Automatic Image Quality Control Off.
Status message, no action required.

XP535 error code P631
Unable to Calibrate.
Status message, no action required.

XP535 error code P632
Automatic Image Quality Control Off.
Developer possibly contaminated or weak. See P633 it's the next error you will
get. See t/s flowchart.

XP535 error code P633
System Near Calibration Limits.
Use MPC Diag to turn on processor Developer pump a couple of times. Run
calibration film to see if error clears. If so, have customer call processor
Service. See t/s flowchart.

969 error code P910
No Communication Between IMS and MCS.
See t/s guide.

969 error code L912/P912
No Clk Received.
Probable output module, cable or PWA-A.

969 error code L913/P913
Parity Error.
Probable output module, cable or PWA-A.

969 error code K1
The laser imager is not communicating.
Keypad error. Try Cycling UKEIB power, see t/s guide.

969 error code K3
The laser imager is not receiving commands correctly.
Keypad error. See t/s guide.

969 error code K4
The laser imager is not transmitting responses correctly.
Keypad error. See t/s guide.

969 error code K9
Communication error.
Keypad error. See t/s guide.

969 error code K58
The images are too large for the format.
Keypad error. See t/s guide.

969 error code K183
The keypad has a stuck key or the footswitch is stuck.
Keypad error. See t/s guide.

969 error code K184
The touch screen is disconnected or damaged.
Keypad error. Check cables inside keypad, replace keypad. See t/s guide.

969 error code K192
Internal keypad error, reset contrast and density settings.
Keypad error. See t/s guide.

969 error code K195
Format select failed.
Keypad error. See t/s guide.


Last Modified June 05, 2000
GE Medical Systems
For GE internal use only
Contact: Server Administrator
-->

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-26 11:21 , Processed in 0.101708 second(s), 32 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表